Leaky Viper Lines

Forums NTA Forum Instruments Leaky Viper Lines

Tagged: 

Viewing 2 posts - 1 through 2 (of 2 total)
  • Author
    Posts
  • #3399
    gpblack
    Moderator

    Has anyone had issues with leaky Viper lines on a Thermo LC? It seems anytime I have to mess with one, it takes me a few tries to get it attached to where it won’t leak. Is this common or do I have worn out fittings (or some kind of user error)?

    #3556
    austinq
    Participant

    We used to have issues with viper and nanoviper lines but now we use a torque screwdriver (H4DETSD) with a hexagonal adaptor bit with a cutout for the line to tighten the lines to ~30 cNm and ~20 cNm for metal viper and glass nanoviper connections respectively. We also do a near fully connected leak check with a pre-detector plug (closed system). This has resolved all viper tightening leak issues for us.